Teledyne T400 Model Vehicle User Manual


 
T400 Ozone Analyzer Operator’s Manual Setup Menu
97
5.4.1.3. RNGE MODE SNGL: Single Range Mode Configuration
The single range mode sets a single maximum range for the both the A1 and A2 analog
outputs. If the single range is selected both outputs are slaved together and will
represent the same reporting range span (e.g. 0-500 ppb), however their electronic signal
levels may be configured for different ranges (e.g. 0-10 VDC vs. 0-.1 VDC; See Section
5.10.1.6).
This Reporting range can b
e set to any value between 0.1 ppb and 10,000 ppb. To select
SINGLE range mode and set the upper limit of the reporting range, press:
SETUP X.X PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SAMPLE RANGE=500.0 PPB O3= XXXX
<TST TST> CAL SETUP
SETUP X.X RANGE MODE MENU
MODE SET UNIT EXIT
SETUP X.X RANGE MODE:SNGL
SNGL DUAL AUTO ENTR EXIT
SETUP X.X RANGE MODE:SNGL
SNGL DUAL AUTO ENTR EXIT
SETUP X.X RANGE MODE MENU
MODE SET UNIT EXIT
SETUP X.X RANGE:500.0 Conc
0 0 5 0 0 .0 ENTR EXIT
Toggle buttons to
select the upper
SPAN limit for the
reporting range
EXIT discards the new
setting
ENTR accepts the
new setting
Note
This is the default reporting range mode for the analyzer.