T400 Ozone Analyzer Operator’s Manual Data Acquisition system (DAS) and APICOM
169
7.4.6. RS-232 REPORT FUNCTION
The DAS can automatically report data to the communications ports, where they can be
captured with a terminal emulation program or simply viewed by the user using the
APICOM software.
To enable automatic COMM port reporting, follow the instruction shown in Section 7.4
then press:
Starting at the EDIT CHANNEL MENU
SETUP X.X 0) CONC: ATIMER 1, 800
PREV NEXT INS DEL EDIT PRNT EXIT
SETUP X.X NAME: CONC
<SET SET> EDIT PRNT EXIT
Continue pressing <SET or SET> until ...
SETUP X.X RS-232 REPORT: OFF
<SET SET> EDIT PRNT EXIT
Use the PREV and
NEXT buttons to
scroll to the DATA
CHANNEL to be
edited
SETUP X.X RS-232 REPORT: OFF
OFF ENTR EXIT
EXIT discards the new
setting
ENTR accepts the
new setting
Toggle these buttons
to turn the RS-232
REPORT feature
ON/OFF