Teledyne T400 Model Vehicle User Manual


 
Teledyne API – Model T400 Photometric Ozone Analyzer Getting Started
71
3.5.2. SETTING THE O
3
GENERATOR LOW-SPAN (MID POINT) OUTPUT
LEVEL
To set the ozone LO SPAN (Midpoint) concentration for the IZS O
3
generator, press:
SETUP X.X 0) DAS_HOLD_OFF=15.0 Minutes
PREV NEXT JUMP EDIT PRNT EXIT
SETUP X.X PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SAMPLE RANGE=500.0 PPB O3= XXXX
<TST TST> CAL SETUP
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG EXIT
SETUP X.X ENTER PASSWORD:818
8 1 8 ENTR EXIT
Continue pressing NEXT until ...
SETUP X.X 4) O3_GEN_LOW1=100.0 PPB
PREV NEXT JUMP EDIT PRNT EXIT
Sets LOW SPAN Point for
RANGE1.
To Set the LOW SPAN point for
RANGE2 in DUAL or AUTO
range modes …
Press NEXT button once more to
select O3_GEN_LOW2
then continue as shown.
SETUP X.X 4) O3_GEN_LOW1=100.0 PPB
0 1 0 0 .0 ENTR EXIT
EXIT discards the new
setting
ENTR accepts the
new setting
Toggle buttonss to change setting
Only Values from 0 to 1500 will be accepted..
A value of 0 turns the lamp OFF.
The ENTR key will disappear if an invalid setting
is attempted.
06870C DCN6332