Teledyne T400 Model Vehicle User Manual


 
T400 Ozone Analyzer Operator’s Manual EPA Protocol Calibration
219
Table 10-2: Activity Matrix for Audit Procedure
AUDIT ACCEPTANCE LIMITS
FREQUENCY AND
METHOD OF
MEASUREMENT
ACTION IF REQUIREMENTS
ARE NOT MET
Multipoint calibration audit
The difference between the
measured and the audit
values as a measure of
accuracy (Sec. 2.0.8 of Ref.
11).
At least once a quarter
(Sec. 2.0.8 of Ref. 11)
Re-calibrate the analyzer.
Data processing audit
Adhere to stepwise
procedure for data reduction
(Sec. 8.4); no difference
exceeding
0.02 ppm.
Perform independent
check on a sample of
recorded data, e.g.,
1 day out of every 2
weeks of data, 2 hours for
each day.
Check all remaining data if
one or more audit checks
exceeds 0.02 ppm.
Systems audit
Method described in this
section of the Handbook.
At the startup of a new
monitoring system, and
periodically as
appropriate; observation
and checklist.
Initiate improved methods
and/or training programs.
06870C DCN6332