Setup Menu T400 Ozone Analyzer Operator’s Manual
104
5.6. SETUP CLK: SETTING THE T400 ANALYZER’S INTERNAL
TIME-OF-DAY CLOCK AND ADJUSTING SPEED
5.6.1.1. Setting the Internal Clock’s Time and Day
The T400 has a time of day clock that supports the DURATION step of the automatic
calibration (ACAL) sequence feature, time of day TEST function, and time stamps on
for the DAS feature and most COMM port messages.
To set the clock’s time and date, press:
SETUP X.X PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SAMPLE RANGE=500.0 PPB O3= XXXX
<TST TST> CAL SETUP
SETUP X.X TIME-OF-DAY CLOCK
TIME DATE EXIT
Toggle HOUR
and/or MINUTE
buttons to
change time.
SETUP X.X TIME: 12:00
1 2 :0 0 ENTREXIT
SETUP X.X TIME: 22:30
2 2:3 0 ENTREXIT
SETUP X.X DATE: 01-JAN-05
0 1JAN0 5 ENTREXIT
SETUP X.X DATE: 18-JUN-05
1 8JUN0 5 ENTREXIT
SETUP X.X TIME-OF-DAY CLOCK
TIME DATE EXIT
Toggle DAY,
MONTH, and/or
YEAR buttons to
change date.
EXIT returns to
SETUP X.X
display
HOUR MINUTE DAY MONTH YEAR
06870C DCN6332