Teledyne T400 Model Vehicle User Manual


 
Getting Started Teledyne API – Model T400 Photometric Ozone Analyzer
34
Verify that there is no apparent external shipping damage. If damage has occurred,
please advise the shipper first, then Teledyne API.
Included with your analyzer is a printed record of the final performance characterization
performed on your instrument at the factory. This record, titled Final Test and
Validation Data Sheet (P/N 04314) is an important quality assurance and calibration
record for this instrument. It should be placed in the quality records file for this
instrument.
With no power to the unit, carefully remove the top cover of the analyzer and check for
internal shipping damage by carrying out the following steps:
1. Remove the setscrew located in the top, center of the Front panel.
2. Remove the two screws fastening the top cover to the unit (one per side towards the
rear).
3. Slide the cover backwards until it clears the analyzer’s front bezel.
4. Lift the cover straight up.
5. Inspect the interior of the instrument to make sure all circuit boards and other
components are in good shape and properly seated.
6. Check the connectors of the various internal wiring harnesses and pneumatic hoses
to make sure they are firmly and properly seated.
7. Verify that all of the optional hardware ordered with the unit has been installed.
These are listed on the paperwork accompanying the analyzer.
3.1.1.1. Ventilation Clearance
Whether the analyzer is set up on a bench or installed into an instrument rack, be sure to
leave sufficient ventilation clearance.
Table 3-1: Ventilation Clearance
AREA MINIMUM REQUIRED CLEARANCE
Back of the instrument
4 in.
Sides of the instrument
1 in.
Above and below the instrument
1 in.
Various rack mount kits are available for this analyzer. See Table 1-1 of this manual for
more information.