Setup Menu T400 Ozone Analyzer Operator’s Manual
114
5.10.1.3. Automatic Group Calibration of the Analog Outputs
To calibrate the outputs as a group with the AOUTS CALIBRATION command, select
the ANALOG I/O CONFIGURATION submenu (see Figure 5-3) then press:
From the
AIO CONFIGURATION SUBMENU
(See Figure 7-4)
Analyzer
automatically
calibrates all
channels for which
AUTO-CAL is turned
ON
If any of the channels have not
been calibrated ot if at least one
channel has AUTO-CAL turned
OFF, this message will read NO.
DIAG AIO NOT AUTO CAL. CONC_OUT_1
DIAG AIO NOT AUTO CAL. CONC_OUT_2
This message
appears when
AUTO-CAL is
Turned OFF for
a channel
DIAG AIO AOUTS CALIBRATED: YES
SET> CAL EXIT
DIAG AIO AOUTS CALIBRATED: NO
SET> CAL EXIT
DIAG ANALOG I/O CONFIGURATION
PREV NEXT ENTR EXIT
DIAG AIO AUTO CALIBRATING TEST_OUTPUT
Note Manual calibration should be used for any analog output set for a 0.1V
output range or in cases where the outputs must be closely matched to
the characteristics of the recording device.