Teledyne T400 Model Vehicle User Manual


 
T400 Ozone Analyzer Operator’s Manual Setup Menu
101
5.5. SETUP PASS: PASSWORD PROTECTION
The menu system provides password protection of the calibration and setup functions to
prevent unauthorized adjustments. When the password feature has been enabled
(SETUP>PASS>ON), the system prompts the user for a password to enter the SETUP
menu. This allows normal operation of the instrument, but requires the password (101)
to access to the menus under SETUP. When PASSWORD is disabled (SETUP>OFF),
any operator can enter the Primary Setup (SETUP) and Secondary Setup
(SETUP>MORE) menus. Whether PASSWORD is enabled or disabled, a password
(default 818) is required to enter the VARS or DIAG menus in the SETUP>MORE
menu.
Table 5-1: Password Levels
PASSWORD LEVEL MENU ACCESS ALLOWED
Null (000) Operation
All functions of the MAIN menu: TEST, GEN, initiate SEQ , MSG, CLR
101
Configuration/Maintenance
Access to primary and secondary SETUP menus when PASSWORD enabled.
818
Configuration/Maintenance
Access to DIAG and VARS menus under the secondary SETUP menu whether
PASSWORD is enabled or disabled.