Setup Menu T400 Ozone Analyzer Operator’s Manual
98
5.4.1.4. RNGE MODE DUAL: Dual Range Mode Configuration
DUAL range mode allows the A1 and A2 outputs to be configured with separate
reporting range spans as well as separate electronic signal levels. The analyzer software
calls these two ranges LOW and HI.
The LOW range setting corresponds with the analog output labeled A1 on the rear
panel of the instrument and is viewable via the test function RANGE1.
The HIGH range setting corresponds with the A2 output and is viewable via the test
function RANGE2.
While the software labels these two ranges as LOW and HI, when in DUAL mode
their upper limits need not conform to that convention. The upper span limit of the
LOW/RANGE1 can be a higher number than that of HI/RANGE2
To set the ranges press following button stroke sequence:
.
SETUP X.X PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SAMPLE RANGE=500.0 PPB O3= XXXX
<TST TST> CAL SETUP
SETUP X.X RANGE MODE MENU
MODE SET UNIT EXIT
SETUP X.X RANGE MODE:SNGL
SNGL DUAL AUTO ENTR EXIT
SETUP X.X RANGE MODE:SNGL
SNGL DUAL AUTO ENTR EXIT
SETUP X.X RANGE MODE MENU
MODE SET UNIT EXIT
SETUP X.X HIGH RANGE:500.0 Conc
0 0 5 0 0 .0 ENTR EXIT
Toggle buttons to
select the upper
SPAN limit for the
reporting range
EXIT discards the new
setting
ENTR accepts the
new setting
SETUP X.X LOW RANGE:500.0 Conc
0 0 5 0 0 .0 ENTR EXIT
The LOW and HIGH
ranges have separate
slopes and offsets for
computing the O
3
concentration.
The two ranges must
be independently
calibrated
.
EXIT discards the
new setting
ENTR accepts the
new setting