Teledyne API Model T201 NH
3
Analyzer Operator Manual Maintenance Schedule
07271B DCN6646 29
4.3 REPLACING THE THERMOCOUPLE
The following instructions provide the necessary information to replace
the existing thermocouple. Figure 4-2 is provided for reference.
You m
ust obtain the following replacement parts kit from TAPI:
010820000 , Type K thermocouple
You will need the following tools:
Nutdriver,
5
/
16
Nutdriver,
11
/
32
Diagonal Cutter
Philips head Screwdriver #2
CAUTION
Avoid damage to the unit: use only Type K thermocouple in the
M501 NH
3
converter. It can be distinguished by its wire colors: red
and yellow. If you have any other thermocouple, do NOT install it;
call TAPI to obtain the correct part.
Figure 4-2. Thermocouple Location