Analyzer Operator Manual Calibration Procedure
07271B DCN6646 21
Table 2-5. Confirming Ammonia Converter Efficiency
Step No. Action Comment
1. Press CAL The T201 enters the calibrate mode from Sample
mode.
2. Channel Selection Press TN
X
then Press ENTR.
3. Range Selection Press LOW range then Press ENTR.
4. Press CONC The following will be displayed:
TNx CONV Exit
Press CONV enter 1.0000, Press ENTR
Press EXIT, EXIT