5.1.7.2. Command Syntax
Commands are not case-sensitive and all arguments within one command (i.e. ID
numbers, keywords, data values, etc.) must be separated with a space character.
All Commands follow the syntax:
X [ID] COMMAND <CR>
Where:
X is the command type (one letter) that defines the type of command.
Allowed designators are listed in Table 5-2 and in Appendix A
[ID] is the ma
chine identification number (Section 5.1.6). Example: the
Command “? 200” follo
wed by a carriage return would print the list of
available commands for the revision of software currently installed in the
instrument assigned ID Number 200.
COMMAND is the command designator: This string is the name of the command being
issued (LIST, ABORT, NAME, EXIT, etc.). Some commands may have
additional arguments that define how the command is to be executed.
Press ? <CR> or refer to Appendix A for a list of available command
designators.
<CR> is a carriage return. All commands must be terminated by a carriage
return (usually achieved by pressing the ENTER key on a computer).
Table 5-3: Teledyne API Serial I/O Command Types
COMMAND COMMAND TYPE
C
Calibration
D
Diagnostic
L
Logon
T
Test measurement
V
Variable
W
Warning
5.1.7.3. Data Types
Data types consist of integers, hexadecimal integers, floating-point numbers, Boolean
expressions and text strings.
Integer data are used to indicate integral quantities such as a number of records, a
filter length, etc. They consist of an optional plus or minus sign, followed by one or
more digits. For example, +1, -12, 123 are all valid integers.
Hexadecimal integer data are used for the same purposes as integers. They
consist of the two characters “0x,” followed by one or more hexadecimal digits (0-9,
A-F, a-f), which is the ‘C’ programming language convention. No plus or minus sign
is permitted. For example, 0x1, 0x12, 0x1234abcd are all valid hexadecimal
integers.
Floating-point numbers are used to specify continuously variable values such as
temperature set points, time intervals, warning limits, voltages, etc. They consist of
an optional plus or minus sign, followed by zero or more digits, an optional decimal
point and zero or more digits. (At least one digit must appear before or after the